    • If the ventilator is not already in diagnostic mode, simultaneously press the ALARM RESET and 100% O2 keys on the front panel for approximately five seconds while turning ventilator power on, then touch OK to enter diagnostic mode. (Service Manual 8-19)
    • Reconnect the oxygen source to the ventilator. (Service Manual 8-19)
  1. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: passo 3, immagine 1 di 1
    • Touch Oxygen on the Hardware screen and set the flow to 1 LPM. Verify that the O2Position display reads from 185 to 525 steps. (Service Manual 8-19)
  2. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: passo 4, immagine 1 di 1
    • Touch Oxygen and set the flow to 180 LPM. Check that the O2 Position display corresponds to the set altitude: (Service Manual 8-19)
    • Connect a patient circuit and analyzer to the ventilator. (Service Manual 8-19)
    • Set the analyzer’s function to read LPM mode and gas select to O2 (set a high flow range capable of measuring 180 LPM oxygen). (Service Manual 8-19)
  3. Philips Respironics V200 Esprit Oxygen Flow Sensor/Valve Replacement: passo 7, immagine 1 di 1
    • Check that set oxygen flow corresponds to measured flow: (Service Manual 8-19/20)
Conclusione
Ethan Norman
Membro da: 05/04/20
157 Reputazione
1 Guida realizzata
Cal Poly, Team S1-G4, Zimmerman Spring 2020 Membro di Cal Poly, Team S1-G4, Zimmerman Spring 2020
CPSU-ZIMMERMAN-S20S1G4
5 Membri
5 Guide realizzate
0 Commenti
Aggiungi Commento
Visualizza Statistiche:
Ultime 24 Ore: 0
Ultimi 7 Giorni: 0
Ultimi 30 Giorni: 4
Tutti i Tempi: 224